LMS Vibration Control: "Transducer on Channel Input2 is too sensitive for current reference profile"

Creator
Creator

I am getting a message in LMS Test.Lab Vibration Control Sine: "Transducer on Channel Input2 is too sensitive for current reference profile". I cannot proceed with my test.

 

I am certain my profile is correct. What should i do?

2 REPLIES

Re: LMS Vibration Control: "Transducer on Channel Input2 is too sensitive for current reference

Siemens Legend Siemens Legend
Siemens Legend

Is it possible that your DAC will be maxed out during your profile with this transducer?  The max DAC output is 10 Volts.

 

For example, if a profile has a max acceleration of 10 g;

 

(10 mV / g accel) * 10 g = 100 mV

(100 mV / g accel) * 10 g = 1 V

(1000 mV / g accel) * 10 g = 10 V

 

The suggestion is to use a less sensitive accelerometer.  In the example above 1000 mV / g would cause the warning but 100 mV/g and 10 mV/g would not.

 

You can also reach out the GTAC support team in North America via the 1-800-955-0000 number in the US or Canada or reach out to us via email.

Re: LMS Vibration Control: "Transducer on Channel Input2 is too sensitive for current reference

Creator
Creator

Went to my calibration lab and got a 10 mV/g accelerometer to replace my 100 mV/g accelerometer.  Everything is working properly now.  Thanks.