Sine Control: Closed Loop Control Parameters are yellow?

Pioneer
Pioneer

Hello,

This question has been on my mind because the wish is to run the shaker to an extreme limit while some of the sine parameters are yellow, velocity and acceleration for example, and i am very hesitant to risk breaking the shaker. But will test lab disable the software from  running the profile while paramters are yellow? 

1 REPLY

Re: Sine Control: Closed Loop Control Parameters are yellow?

Siemens Valued Contributor Siemens Valued Contributor
Siemens Valued Contributor

The Reference Profile is checked against the shaker parameters as follows:

Parameters list

Max. acceleration

This shows the maximum acceleration (zero to peak) associated with the profile. It is checked against the shaker parameters. An orange light indicates that the value exceeds the limitation of the current shaker.

Max. velocity

This shows the maximum velocity (zero to peak) associated with the profile. It is checked against the shaker parameters. An orange light indicates that the value exceeds the limitation of the current shaker.

Max. displacement

This shows the maximum displacement (zero to peak) associated with the profile. It is checked against the shaker parameters. An orange light indicates that the value exceeds the limitation of the selected shaker.

 

A Yellow indicator will allow you to continue to test, but you are near the limits of the shaker parameters entered into the Shaker Definition file.  If you see a red indicator you will not be able to continue until you change the reference profile.